KHD Titration Error

Users Who Are Viewing This Thread (Total: 1, Members: 0, Guests: 1)

Crimsonvice

Valuable Member
View Badges
Joined
Oct 12, 2015
Messages
1,375
Reaction score
948
Location
Cincinnati
Rating - 0%
0   0   0
I have a KHD that was running fine for two weeks and then my alk readings started to increase and finally it started to give me "Error: Titration takes too long."

I was trying to troubleshoot it with Vinny, but the company is on vacation for the next two weeks. Does anyone have any suggestions? I already re-calibrated pH probe and pumps, vented reagent tube, and increased my sample to 90 mls.
 

Dan_P

5000 Club Member
View Badges
Joined
Sep 21, 2018
Messages
6,672
Reaction score
7,163
Rating - 0%
0   0   0
I have a KHD that was running fine for two weeks and then my alk readings started to increase and finally it started to give me "Error: Titration takes too long."

I was trying to troubleshoot it with Vinny, but the company is on vacation for the next two weeks. Does anyone have any suggestions? I already re-calibrated pH probe and pumps, vented reagent tube, and increased my sample to 90 mls.

Can you determine your system’s alkalinity with a test kit?
 

road_runner

Valuable Member
View Badges
Joined
Jan 6, 2017
Messages
2,325
Reaction score
2,293
Rating - 0%
0   0   0
I have a KHD that was running fine for two weeks and then my alk readings started to increase and finally it started to give me "Error: Titration takes too long."

I was trying to troubleshoot it with Vinny, but the company is on vacation for the next two weeks. Does anyone have any suggestions? I already re-calibrated pH probe and pumps, vented reagent tube, and increased my sample to 90 mls.
This means reagent is not going through the device at one point or another. Did you check if there is any bubble in the tubing?
Do you see reagent going through the tubing?
Do you hear the titration peristaltic pump working and routating?
Basically go through the reagent path from start to end and see if there is disconnection, bubble or blockage..
 

WWIII

2500 Club Member
View Badges
Joined
Jan 13, 2016
Messages
3,739
Reaction score
7,701
Location
Louisville, KY
Rating - 0%
0   0   0
Does the ph value drop down while the test is happening? Should be able to watch it go down to under 4.5 then it will give the result.

Are you sure the doser heads are all operating flawlessly and calibrated correctly? No blue chips or sticking.

Have you checked for and removed the tiny inline filter inside?

@Ohashimz nailed it. Not getting enough acid some way? some how.
 
OP
OP
Crimsonvice

Crimsonvice

Valuable Member
View Badges
Joined
Oct 12, 2015
Messages
1,375
Reaction score
948
Location
Cincinnati
Rating - 0%
0   0   0
This means reagent is not going through the device at one point or another. Did you check if there is any bubble in the tubing?
Do you see reagent going through the tubing?
Do you hear the titration peristaltic pump working and routating?
Basically go through the reagent path from start to end and see if there is disconnection, bubble or blockage..

It is definitely going through the tube. I have had a GHL Doser for years, so I am pretty comfortable with calibrating, so I do not think that is the issue. Vinny suggested an air bubble too, but I am not seeing one in my line.

Does the ph value drop down while the test is happening? Should be able to watch it go down to under 4.5 then it will give the result.

Are you sure the doser heads are all operating flawlessly and calibrated correctly? No blue chips or sticking.

Have you checked for and removed the tiny inline filter inside?

@Ohashimz nailed it. Not getting enough acid some way? some how.

I am not sure about the filter. Would this involve poping open the top?

I watched the pH value, and it dropped to about 5.62 and then started rising again. So like you guys said, it seems to not be getting enough acid for the test. I have the dosing pump set to a speed of 0. Is this correct?
 

road_runner

Valuable Member
View Badges
Joined
Jan 6, 2017
Messages
2,325
Reaction score
2,293
Rating - 0%
0   0   0
It is definitely going through the tube. I have had a GHL Doser for years, so I am pretty comfortable with calibrating, so I do not think that is the issue. Vinny suggested an air bubble too, but I am not seeing one in my line.



I am not sure about the filter. Would this involve poping open the top?

I watched the pH value, and it dropped to about 5.62 and then started rising again. So like you guys said, it seems to not be getting enough acid for the test. I have the dosing pump set to a speed of 0. Is this correct?
If you seeing ph drop that means titration is happening for sure...bizzar..
Sorry am not enough help. Will follow hopefully you figure it out. Ghl support is great they will figure it out.
 

Ditto

Valuable Member
View Badges
Joined
Apr 25, 2017
Messages
2,229
Reaction score
2,972
Location
Albany, NY
Rating - 0%
0   0   0
If you open up the unit, and in the GHL app for the KH director diagnostic section if you enable diagnostic and hit the pinch valve do you hear it?

I would pull the tube out from the pinch valve and reseat it into the pinch valve.

Next hit the vent reagent line and see the results foes the reagent flush?

I had a issue with the filter on the Reagent line get clogged you will see air bubbles inside the reagent line to the pinch valve. It may also popped out, i am not sure if they are on the newer units, mine was a first generation run.

The 4 screws are on the bottom, empty the chamber cost before taking it apart. Then leave the cover off and run the test.
 

Ditto

Valuable Member
View Badges
Joined
Apr 25, 2017
Messages
2,229
Reaction score
2,972
Location
Albany, NY
Rating - 0%
0   0   0
Vent the Reagent a few times also and run a test does it say it or does it go off the scale also.
 

WWIII

2500 Club Member
View Badges
Joined
Jan 13, 2016
Messages
3,739
Reaction score
7,701
Location
Louisville, KY
Rating - 0%
0   0   0
Speed 0 is correct.

@Ditto has the rest covered as far as what you can do to diagnose. When I bought an older khd that someone had sitting around I had to reseat the tubing in the pinch valve as it had stuck itself shut. Once I did that it was fine (past 6 months). Then there is a tiny white filter in the line on the inside of the unit (yes pop the top actually the bottom) check for that and remove it as it's not necessary and not included on the current production units.

I'd do those things with the pinch valve line and check for filter. Then run the other diagnostic tests ditto describes. We can go from there.
 
OP
OP
Crimsonvice

Crimsonvice

Valuable Member
View Badges
Joined
Oct 12, 2015
Messages
1,375
Reaction score
948
Location
Cincinnati
Rating - 0%
0   0   0
Okay, I popped the top and checked the pinch valve. Everything seems to be functioning okay. I do not see an internal filter.

dd8791a49a2395345774e752e829f782.jpg


31e28643ab8c198336087ff98094bd19.jpg


d64eea94b4e60b80bfb9fd3e452b454c.jpg






Thoughts?
 
OP
OP
Crimsonvice

Crimsonvice

Valuable Member
View Badges
Joined
Oct 12, 2015
Messages
1,375
Reaction score
948
Location
Cincinnati
Rating - 0%
0   0   0
Well I finally got it to at least measure something. I replaced the reagent 1/4" tube from a stiff orange line to a soft vinyl. Maybe this did the trick. The orange line wasn't clogged or anything, so I don't know what the difference is.

685a5d2ae33eabf364a8fd522e3a80f9.jpg
 

road_runner

Valuable Member
View Badges
Joined
Jan 6, 2017
Messages
2,325
Reaction score
2,293
Rating - 0%
0   0   0
Well I finally got it to at least measure something. I replaced the reagent 1/4" tube from a stiff orange line to a soft vinyl. Maybe this did the trick. The orange line wasn't clogged or anything, so I don't know what the difference is.

685a5d2ae33eabf364a8fd522e3a80f9.jpg
Glad it worked out. Still talk to ghl and see what they say and post it here pls so we all learn.
 

WWIII

2500 Club Member
View Badges
Joined
Jan 13, 2016
Messages
3,739
Reaction score
7,701
Location
Louisville, KY
Rating - 0%
0   0   0
Well I finally got it to at least measure something. I replaced the reagent 1/4" tube from a stiff orange line to a soft vinyl. Maybe this did the trick. The orange line wasn't clogged or anything, so I don't know what the difference is.

685a5d2ae33eabf364a8fd522e3a80f9.jpg


Yeah I am stumped. I'm not sure why the rigid ro tubing would make a difference unless somehow it was letting air in and/or losing prime in between tests? Maybe the inside diameter is different, but the dosing head calibration should account for any of that. How many ml/min is the reagent head calibrated to?
 

kodo28

Active Member
View Badges
Joined
May 27, 2016
Messages
289
Reaction score
172
Rating - 0%
0   0   0
I could be wrong but the inline filter is not just next to the connector on your picture ? looks like the filter as drifted to the connector.

1563891346255.png



This is what look like mine on the connector side

1563891437648.png
 
Last edited:
OP
OP
Crimsonvice

Crimsonvice

Valuable Member
View Badges
Joined
Oct 12, 2015
Messages
1,375
Reaction score
948
Location
Cincinnati
Rating - 0%
0   0   0
I could be wrong but the inline filter is not just next to the connector on your picture ? looks like the filter as drifted to the connector.

1563891346255.png



This is what look like mine on the connector side

1563891437648.png
Ok so that filter is what needs to be removed?
 

kodo28

Active Member
View Badges
Joined
May 27, 2016
Messages
289
Reaction score
172
Rating - 0%
0   0   0
Ok so that filter is what needs to be removed?

Yes it should be removed ;)
 

Vinny@GHLUSA

Valuable Member
View Badges
Joined
Jul 27, 2016
Messages
1,149
Reaction score
1,509
Rating - 0%
0   0   0
As @kodo28 mentioned, that in-line filter should be removed. It looks like it became clogged and could no longer allow air to flow through the tube. After enough tests, that filter will make its way to the end of the vent tube and restrict even more air...ultimately leading to the titration error.

Remove that filter and keep the original internal reagent tube in-place.
 

Rock solid aquascape: Does the weight of the rocks in your aquascape matter?

  • The weight of the rocks is a key factor.

    Votes: 10 8.1%
  • The weight of the rocks is one of many factors.

    Votes: 43 34.7%
  • The weight of the rocks is a minor factor.

    Votes: 39 31.5%
  • The weight of the rocks is not a factor.

    Votes: 31 25.0%
  • Other.

    Votes: 1 0.8%
Back
Top