RODI Unit Reading

TheSeas

Valuable Member
View Badges
Joined
Jan 24, 2021
Messages
1,249
Reaction score
1,483
Location
NJ
Rating - 0%
0   0   0
Hi quick question, I recently purchased a BRS 4 stage RODI unit. In viewing the TDS in and out I am reading 1 in and 1 out. This is after flushing the waste line, using for over 2hrs and checking the senor probes. Is this common or is there something wrong with the unit. I contacted BRS and they stated it is fine. I appreciate your thoughts.
 
Last edited:

Uncle99

7500 Club Member
View Badges
Joined
Jul 22, 2018
Messages
8,959
Reaction score
13,169
Location
Province of Ontario
Rating - 0%
0   0   0
1 out after DI stage would be normal.
1 in, (from source) would be outstanding, but this number seems off, it should be maybe 100-300ppm TDS out of the tap, down to say 3-10ppm after RO, and 0-1ppm after DI.
 

KStatefan

2500 Club Member
View Badges
Joined
Jun 24, 2015
Messages
4,119
Reaction score
3,996
Location
MHK
Rating - 0%
0   0   0
If you are getting 1 TDS in to the di resin there is no reason it should not be 0 TDS out. That is assuming the TDS probes are before and after the DI.
 
OP
OP
TheSeas

TheSeas

Valuable Member
View Badges
Joined
Jan 24, 2021
Messages
1,249
Reaction score
1,483
Location
NJ
Rating - 0%
0   0   0
The probes are before and after. Something is off which is not making me feel comfortable trusting the readings.
 

Being sticky and staying connected: Have you used any reef-safe glue?

  • I have used reef safe glue.

    Votes: 119 88.1%
  • I haven’t used reef safe glue, but plan to in the future.

    Votes: 8 5.9%
  • I have no interest in using reef safe glue.

    Votes: 5 3.7%
  • Other.

    Votes: 3 2.2%
Back
Top