Using the actual equations you can see the precision problem clearly with this method.
Let's assume an actual CO2 measurement of 415 and pH of 8.10
If the sensor has a +/- 0.01 repeatability for pH, that would result in Alk readings of 7.54 to 7.97 dKH.
If you add in just a +/- 1% CO2 reading you will get 7.45 to 8.06 dKH readings or roughly 0.6 dKH difference.
Many would find this to be too inaccurate for their reefing needs.