I bought an ATI ICP test for the first time. Does anyone know if outside temperatures when mailing them off affect the water samples?
Last edited:
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
Most likely it will, but what we don’t know is a) how much it will affect the sample, and b) whether it will matter to the decisions you make about aquarium maintenance.I bought an ATI ICP test for the first time. Does anyone know if outside temperatures when mailing them off affect the water samples?
It will be interesting to compare the two, although here is where my ranting comes in. If we don’t know what the standard deviation is in the measurement, the 95% confidence interval is another term you might see, how do we know that any difference we see is just noise in the data or a real difference in the two setups?Thanks @Lasse and @Randy Holmes-Farley
@Dan_P I bought the icp tests to compare the water between my two tanks. One has coraline growing out the wazoo and the other only gets it on the powerheads and overflow box. They have the same type of rock, maintenance, same source water and salt mix. Only differences are size (90 vs 29 gallons) and lighting. Chinese black box vs lumi lite pro.
I’m mostly just curious about the differences between the two for all the stuff I can’t test for.
This means that we must stop all testes we do with start with all hobby tests. As I have seen - there is no method description, no interference tables or no information at all in most hobby tests . There is devices with a accuracy of ± 0.02 that we use in order to withhold total concentration around 0.03 ppm. Devices that says to be useful for use by not analytic trained persons and give valid results in such a low concentrations that even professional labs have difficulties with it.Most likely it will, but what we don’t know is a) how much it will affect the sample, and b) whether it will matter to the decisions you make about aquarium maintenance.
See aboveIf we don’t know what the standard deviation is in the measurement, the 95% confidence interval is another term you might see, how do we know that any difference we see is just noise in the data or a real difference in the two setups?
This means that we must stop all testes we do with start with all hobby tests. As I have seen - there is no method description, no interference tables or no information at all in most hobby tests . There is devices with a accuracy of ± 0.02 that we use in order to withhold total concentration around 0.03 ppm. Devices that says to be useful for use by not analytic trained persons and give valid results in such a low concentrations that even professional labs have difficulties with it.
If we should talk about reliability in testing and forget the elephant in the room - the hobby tests and all the money going that way. As an example - the nitrite interference in nitrate tests - how many nitrat tests give a warning for that, especially in the low range?
See above
I have done - during the last 4 years - 17 ICP test of my own aquarium and at my work we have done +200 tests in different aquariums. These results have been very useful in order to tune in the aquariums and help checking our own measurements of the 4 "big". Both at home and at work we have achieved success during a long period without any serve crashes. When an aquarium is tuned in - can take a year or two - the results from test to test does not differ very much if you not actively dose things.
Of cause - you can't rely on one single test - these tests - lika all others we use - its the trend that is important. If we can get people to understand that a test - even if the result is 100% correct - is only a snapshot of the picture at the test moment it will be more helpful compared with rants about things we really do not know about.
ICP test have help me to hinder a crash in my aquarium. My Hanna shows 0.1 ppm - my Triton test show 0,018 ppm. This was not a issue only for me. R2R was full of threads speculation that the P content in the sample was altered during transportation and that Hanna Checker testing was the most accurate method to rely on. However - all my other observations shows the opposite - I decide to start dosing PO4 in spite of the fact that my Hanna rise to around 0.14. Meaningless to say but all things turned to the better after this change - everything indicate that the ICP test was more near the real concentration compared with my Hi-774. After this - I have purchase a Red Sea Pro PO4 test and even this indicate a concentration around 0.08 ppm lower than the Hi-774
This thread will be interesting to follow - in post #92 Ehsan from Triton steps in
Sincerely Lasse
There is daily tests taken of PO4 and alkalinity in most of the systems. Salifert and Hanna Hi 713. ICP test every 2 months as a checkup. We use calcium reactors and Triton Core 7 - both are balanced and we need only to analyse alkalinity on a daily base. @Sallstrom knows more of this. Both alkalinity and PO4 can change quickly in these heavy coral populated tanksThanks @Lasse. In the aquariums that you work with (200+ tests), are regular tests performed with Red Sea, Hanna checkers etc. also or do they solely use ICP? An ICP test seems more convenient and a lot easier in that situation.
This means that we must stop all testes we do with start with all hobby tests. As I have seen - there is no method description, no interference tables or no information at all in most hobby tests .
There is devices with a accuracy of ± 0.02 that we use in order to withhold total concentration around 0.03 ppm. Devices that says to be useful for use by not analytic trained persons and give valid results in such a low concentrations that even professional labs have difficulties with it.
If we should talk about reliability in testing and forget the elephant in the room - the hobby tests and all the money going that way. As an example - the nitrite interference in nitrate tests - how many nitrat tests give a warning for that, especially in the low range?
I have done - during the last 4 years - 17 ICP test of my own aquarium and at my work we have done +200 tests in different aquariums. These results have been very useful in order to tune in the aquariums and help checking our own measurements of the 4 "big". Both at home and at work we have achieved success during a long period without any serve crashes. When an aquarium is tuned in - can take a year or two - the results from test to test does not differ very much if you not actively dose things.
Of cause - you can't rely on one single test - these tests - lika all others we use - its the trend that is important. If we can get people to understand that a test - even if the result is 100% correct - is only a snapshot of the picture at the test moment it will be more helpful compared with rants about things we really do not know about.
ICP test have help me to hinder a crash in my aquarium. My Hanna shows 0.1 ppm - my Triton test show 0,018 ppm. This was not a issue only for me. R2R was full of threads speculation that the P content in the sample was altered during transportation and that Hanna Checker testing was the most accurate method to rely on. However - all my other observations shows the opposite - I decide to start dosing PO4 in spite of the fact that my Hanna rise to around 0.14. Meaningless to say but all things turned to the better after this change - everything indicate that the ICP test was more near the real concentration compared with my Hi-774. After this - I have purchase a Red Sea Pro PO4 test and even this indicate a concentration around 0.08 ppm lower than the Hi-774
@Rick Mathew We are inline with each other. You have found ways around the limitation of the hobby tests and you are able to get results that you can relay on. But you are a skill statistician and know your math. 99 % of people using these test do not have your skill or my skepticism to my own technique. The relay on the figures and make decisions that can be catastrophic. i can´t count all threads that start - high ammonia readings what to do. after that all answers - change water, do that or do that - your fish is going to die. Reading the thread often shows up a total ammonia reading of 0.2 - 0.25 ppm. The truth of this reading is that many total ammonium hobby test often show 0.25 even if it zero. and if the result was the right - it is no danger because a reading of 0.25 NH3/NH4 only will give a concentration of 0.02ppm of the toxic form NH3 at pH 8.2 - and its a safe value.
The Hi 774 is basically the same as you use but convert the reading into PO4 directly. And I have worked out a procedure there I analyse the same sample 10 times in a row and after that can take an average - it means that I with rather high accuracy comes around the ±0.02 limit. But - in my case - it shows that my meter probably have a systematic error of 0.08 ppm and therefore I - as you - use my outside tests to get a reading that I rely on. I was ranting in my last post because there is a tendency that when outside tests and hobby tests show different - it is always the outside tests that you can´t rely on. For me - the outside test is the ones I adapt my use of hobby tests too. I know that my Ca readings always was - as yours 15 - 20 ppm lower compared with the ICP test. After seen this for many tests - I adapt may way of reading my hobby test and now - they are nearly the same.
One thing is the nitrate tests. I have had contact with one of the ICP testers here in Europe that the same time do advanced NO3 and NO2 tests. Their database show that it is very common with NO2 levels around 0.02-0.05 ppm - but there is aquariums with zero upt to 0.15 NO2. Already levels around 0.02 - 0.05 will give NO3 readings below 5 ppm a huge error.
Today - I use Tritons N-DOC in order to know what maximal my NO3 can be. And if I am below max 10 ppm - I fine with that because it probably indicate that my NO3 levels is around 5.
Sincerely Lasse
@Rick Mathew We are inline with each other. You have found ways around the limitation of the hobby tests and you are able to get results that you can relay on. But you are a skill statistician and know your math. 99 % of people using these test do not have your skill or my scepticism to my own technique. The relay on the figures and make decisions that can be catastrophic. i can´t count all threads that start - high ammonia readings what to do. after that all answers - change water, do that or do that - your fish is going to die. Reading the thread often shows up a total ammonia reading of 0.2 - 0.25 ppm. The truth of this reading is that many total ammonium hobby test often show 0.25 even if it zero. and if the result was the right - it is no danger because a reading of 0.25 NH3/NH4 only will give a concentration of 0.02ppm of the toxic form NH3 at pH 8.2 - and its a safe value.
The Hi 774 is basically the same as you use but convert the reading into PO4 directly. And I have worked out a procedure there I analyse the same sample 10 times in a row and after that can take an average - it means that I with rather high accuracy comes around the ±0.02 limit. But - in my case - it shows that my meter probably have a systematic error of 0.08 ppm and therefore I - as you - use my outside tests to get a reading that I rely on. I was ranting in my last post because there is a tendency that when outside tests and hobby tests show different - it is always the outside tests that you can´t rely on. For me - the outside test is the ones I adapt my use of hobby tests too. I know that my Ca readings always was - as yours 15 - 20 ppm lower compared with the ICP test. After seen this for many tests - I adapt may way of reading my hobby test and now - they are nearly the same.
One thing is the nitrate tests. I have had contact with one of the ICP testers here in Europe that the same time do advanced NO3 and NO2 tests. Their database show that it is very common with NO2 levels around 0.02-0.05 ppm - but there is aquariums with zero upt to 0.15 NO2. Already levels around 0.02 - 0.05 will give NO3 readings below 5 ppm a huge error.
Today - I use Tritons N-DOC in order to know what maximal my NO3 can be. And if I am below max 10 ppm - I fine with that because it probably indicate that my NO3 levels is around 5.
Sincerely Lasse
HI736 | HI-774 | DIFFERENCE |
0.05 | 0.08 | 0.03 |
0.07 | 0.09 | 0.02 |
0.09 | 0.12 | 0.03 |
0.06 | 0.08 | 0.02 |
0.08 | 0.09 | 0.01 |
0.06 | 0.06 | 0 |
0.06 | 0.08 | 0.02 |
0.07 | 0.09 | 0.02 |
0.07 | 0.09 | 0.02 |
0.06 | 0.09 | 0.03 |
0.02 | 0.05 | 0.03 |
0.04 | 0.08 | 0.04 |
0.05 | 0.05 | 0 |
0.03 | 0.05 | 0.02 |
0.05 | 0.07 | 0.02 |
0.05 | 0.07 | 0.02 |
0.03 | 0.05 | 0.02 |
0.04 | 0.06 | 0.02 |
0.04 | 0.05 | 0.01 |
0.04 | 0.05 | 0.01 |
0.03 | 0.04 | 0.01 |
0.04 | 0.08 | 0.04 |
0.05 | 0.06 | 0.01 |
0.04 | 0.06 | 0.02 |
0.04 | 0.04 | 0 |
AVG | 0.02 |
Lasse: Below is some (not all) of the data I have on comparing my measurements of PO4 on the HI-736 and HI-774. The tests were done on the same water samples taking at least 3 measurements and getting the average...As you can see my HI-774 measures (on Average) .02 higher than my Hi-736...Have no idea why ... More to explore!...so I just thought I would share it with you.
HI736 HI-774 DIFFERENCE 0.05 0.08 0.03 0.07 0.09 0.02 0.09 0.12 0.03 0.06 0.08 0.02 0.08 0.09 0.01 0.06 0.06 0 0.06 0.08 0.02 0.07 0.09 0.02 0.07 0.09 0.02 0.06 0.09 0.03 0.02 0.05 0.03 0.04 0.08 0.04 0.05 0.05 0 0.03 0.05 0.02 0.05 0.07 0.02 0.05 0.07 0.02 0.03 0.05 0.02 0.04 0.06 0.02 0.04 0.05 0.01 0.04 0.05 0.01 0.03 0.04 0.01 0.04 0.08 0.04 0.05 0.06 0.01 0.04 0.06 0.02 0.04 0.04 0 AVG 0.02
Now that you mention it I seem to recall a discussion around forcing the slope to go through the zero. I think it was in the development of the nitrate test using the 736 I'll have to go back and look and see....... good memoryCould the 0.02 ppm shift be a result of forcing one of the Hanna Checker regression lines through zero? Didn’t @taricha find something odd about the near zero readings in the ULR P Checker?
I was also thinking that Hanna might have forced a regression through zero forone of their Checkers and produced the 0.02 difference between their two Checkers. It might require a spectroscopy study to investigate.Now that you mention it I seem to recall a discussion around forcing the slope to go through the zero. I think it was in the development of the nitrate test using the 736 I'll have to go back and look and see....... good memory
I was also thinking that Hanna might have forced a regression through zero forone of their Checkers and produced the 0.02 difference between their two Checkers. It might require a spectroscopy study to investigate.
What I noticed was that the amount of color required above zero in order to get a reading is more than the color difference error at other concentrations. It requires about 2-3 times the normal uncertainty to get a reading above zero. In other words, they built in a buffer against false positives.Could the 0.02 ppm shift be a result of forcing one of the Hanna Checker regression lines through zero? Didn’t @taricha find something odd about the near zero readings in the ULR P Checker?
If you want to probe that for your checkers, then use a constant color solution - like drops of tea or koolaid or something, don't bother with the hassle of a chemical test just to check the low range color linearity.Could you find it using standards from...say .01 to .04 measured on both...and plot the slope. ..or am I confused