Hannah Alkalinity Refill inaccurate.

75pxatr

Community Member
View Badges
Joined
Mar 12, 2011
Messages
35
Reaction score
27
Location
poland oh
Rating - 0%
0   0   0
I have had the same problem with Hannah twice in the last 4 years. Now I will only use Salifert and the Hannah as a double check if necessary. I would never trust Hannah as my sole method of determining alk.
 

Justfebreezeit

Valuable Member
View Badges
Joined
Nov 30, 2017
Messages
1,489
Reaction score
1,789
Location
SoFlo
Rating - 0%
0   0   0
My new bottle that tested 1 dkh lower out of the gate seems to be consistent to itself at least. Getting normal the normal drop between dosing. Honestly this is fine for me. I hate titration.
 

Retfitter

New Member
View Badges
Joined
Mar 28, 2017
Messages
12
Reaction score
5
Location
Pocatello, Idaho
Rating - 0%
0   0   0
I have had this issue with 1 bottle of reagent. Since then, whenever I open a new bottle, I compare it with another brand of test.
 

DivingTheWorld

Acroholic
View Badges
Joined
Sep 6, 2018
Messages
2,396
Reaction score
7,375
Location
NorCal
Rating - 0%
0   0   0
so is your livestock suffering?

Mine have, yes. I've been running in mid 7's (Hanna) and recently allowed it to increase to just above 8 dkh since I figured this was a safer range and closer to what many acro tanks run. My acros did not like it and I subsequently dropped it back down. My test today was 7.784 Hanna, 8.6 Salifert. I would imagine when I was up around 8.2 Hanna, I was probably in reality above 9 dkh. Our tanks are all about consistency, but trying to maintain consistency with a test kit that is 1 dkh off is a very tricky proposition.
 

Hanna Instruments

Active Member
View Badges
Joined
Feb 16, 2017
Messages
373
Reaction score
460
Location
Woonsocket, Rhode Island
Rating - 0%
0   0   0
Always rinse your cuvettes with RODI water after each test and with saltwater sample before a new test is preformed to prevent unintended dilution of sample.


•The alkalinity reagent is very pH sensitive and any saltwater or outside contaminants can alter the pH of the reagent causing inaccurate results. Make sure no saltwater or other contaminants enter your reagent bottle and if questionable results are obtained open a fresh reagent and preform multiple tests to confirm results.


•Always use the new syringe and tip provided with each new reagent. Do not use the old syringes/tips for new reagent bottles.


•Using a syringe with dried or crusty reagent on its tip is not ideal and syringes should be cleaned with RODI water periodically to prevent corrosion.
 

DivingTheWorld

Acroholic
View Badges
Joined
Sep 6, 2018
Messages
2,396
Reaction score
7,375
Location
NorCal
Rating - 0%
0   0   0
I’m struggling to find something of value in that response. Speaking for myself, I already do all of that. The issue is not the procedure, it’s the reagent. I hate to say it, but I may just have to switch to Salifert.
 

Hanna Instruments

Active Member
View Badges
Joined
Feb 16, 2017
Messages
373
Reaction score
460
Location
Woonsocket, Rhode Island
Rating - 0%
0   0   0
I’m struggling to find something of value in that response. Speaking for myself, I already do all of that. The issue is not the procedure, it’s the reagent. I hate to say it, but I may just have to switch to Salifert.
Our HI772 dKH Alkalinity Checker has an accuracy statement of ±0.3 dKH ±5% of reading. So if your Checker reads 10.0 dKH the measure of uncertainty would be ±0.5 dKH or a range from 9.5-10.5 dKH

The major advantage of using a colorimeter, or digital test, is you can measure accuracy without subjective human error (so long as instructions are followed). Other chemical test kit brands may claim to have an accuracy statements but you cannot quantify inaccuracies brought about by judging color changes. We regularly test our lot numbers from stock against high end lab equipment used in professional water quality analysis centers. Our Checkers/reagents always preform within, or better, than their accuracy statement.

We do not use accuracy statements for any chemical test kits, nor do most analytical chemistry brands. This is because an accuracy statement is usually reserved for instrumentation with digital displays of some sort. Human interpretation of color is subjective and thus you cannot measurably quantify what that measure of uncertainty will be due to differences in how people observe colors.
 

DivingTheWorld

Acroholic
View Badges
Joined
Sep 6, 2018
Messages
2,396
Reaction score
7,375
Location
NorCal
Rating - 0%
0   0   0
Our HI772 dKH Alkalinity Checker has an accuracy statement of ±0.3 dKH ±5% of reading. So if your Checker reads 10.0 dKH the measure of uncertainty would be ±0.5 dKH or a range from 9.5-10.5 dKH

The major advantage of using a colorimeter, or digital test, is you can measure accuracy without subjective human error (so long as instructions are followed). Other chemical test kit brands may claim to have an accuracy statements but you cannot quantify inaccuracies brought about by judging color changes. We regularly test our lot numbers from stock against high end lab equipment used in professional water quality analysis centers. Our Checkers/reagents always preform within, or better, than their accuracy statement.

We do not use accuracy statements for any chemical test kits, nor do most analytical chemistry brands. This is because an accuracy statement is usually reserved for instrumentation with digital displays of some sort. Human interpretation of color is subjective and thus you cannot measurably quantify what that measure of uncertainty will be due to differences in how people observe colors.

Wow, +/- 0.5 dKH.
 

Danj

Community Member
View Badges
Joined
Apr 19, 2014
Messages
99
Reaction score
116
Location
Atlantic Highlands, NJ
Rating - 0%
0   0   0
Thanks Hanna, appreciate you weighing in on the discussion and I definitely agree, digital displays have significant advantages over subjective color comparisons. I do have a question regarding the specs, is there a published repeatability rating of the device? I think this would go a long way towards aligning with the mantra that you see in our hobby, vaguely summarized as "I don't care too much about absolute numbers as long as they're stable."

3989c78c08959287d88add6963a648a7.jpg


(Image provided just to nip any terminology confusion in the bud, since I know there's often a lot of confusion about what accuracy, precision, and repeatability mean in a scientific context)


If the gauge repeatibilty is low, then I think it has a tendency to send many of us into a mad scientist loop where we're correcting for perceived fluctuations that don't exist.


The impression I'm getting is that we can expect individual batches to be highly repeatable test-to-test within the batch but with the specified accuracy offset you've stated. Basically the leftmost target in my image. Then when a new batch is introduced, it would have a roughly equally tight grouping with a new offset as dictated by the accuracy spec. Would you say I have that summarized fairly?
 

Scorpius

2500 Club Member
View Badges
Joined
Mar 6, 2017
Messages
3,660
Reaction score
3,754
Location
Indiana
Rating - 0%
0   0   0
Hannah, I love your ULR Phosphorus Checker, but for everything else I've switched to Salifert.
 
OP
OP
Terry Brown

Terry Brown

Community Member
View Badges
Joined
Oct 30, 2017
Messages
60
Reaction score
95
Rating - 0%
0   0   0
I’m struggling to find something of value in that response. Speaking for myself, I already do all of that. The issue is not the procedure, it’s the reagent. I hate to say it, but I may just have to switch to Salifert.

DivingTheWorld, I agree with you. I am very through with my testing procedures. And all I heard Hanna saying is their detectors are accurate. Their statements do not address the consistency of the Reagent.

I have many thousands of dollars invested in Corals and fish. Hannah checkers being 1 DKH off, as many people have stated in this thread, is not acceptable.

I am going to have to go back to the Salifert.
 

Danj

Community Member
View Badges
Joined
Apr 19, 2014
Messages
99
Reaction score
116
Location
Atlantic Highlands, NJ
Rating - 0%
0   0   0
Just to play devil's advocate, and because I can't Google it at the moment. Does salifert publish an accuracy spec? How does it compare?
 

foxt

Valuable Member
View Badges
Joined
Apr 6, 2016
Messages
1,633
Reaction score
2,363
Rating - 0%
0   0   0
I was searching R2R for something else related to alk, and came across this thread. I think the OP is on to something with respect to inconsistency between reagent lots. I just switched bottles from lot #7522 expiring in 11/2019 to a lot that expires in 2020 (I forget the lot # and am too lazy at the moment to go look). I found a difference of 0.8dKH between the lots - taking samples back-to-back in the same vials, same process, etc. I did it twice with the new lot to be sure (could not retest the old lot because the bottle is now empty).

I've never noticed such a big swing when starting a new bottle before, and I was hoping that Hanna would directly address this with either a comment about a bad set of lots, or at least to tell us that this is "normal" so that I would know to always test the new lot side by side with the old lot.

Now I'm going away for two days, and will return with a salifert test kit on Saturday to see how it compares to this new lot in the Hanna. Like others, I switched to Hanna because I liked the readout vs. trying to interpret colors, but if there is a lack of reliability on the Hanna reagents, then I may be switching back.
 
Last edited:

zoasaholic

Well-Known Member
View Badges
Joined
Aug 10, 2013
Messages
987
Reaction score
563
Location
Bay Area,CA
Rating - 0%
0   0   0
I was searching R2R for something else related to alk, and came across this thread. I think the OP is on to something with respect to inconsistency between reagent lots. I just switched bottles from lot #7522 expiring in 11/2019 to a lot that expires in 2020 (I forget the lot # and am too lazy at the moment to go look). I found a difference of 0.8dKH between the lots - taking samples back-to-back in the same vials, same process, etc. I did it twice with the new lot to be sure (could not retest the old lot because the bottle is now empty).

I've never noticed such a big swing when starting a new bottle before, and I was hoping that Hanna would directly address this with either a comment about a bad set of lots, or at least to tell us that this is "normal" so that I would know to always test the new lot side by side with the old lot.

Now I'm going away for two days, and will return with a salifert test kit on Saturday to see how it compares to this new lot in the Hanna. Like others, I switched to Hanna because I liked the readout vs. trying to interpret colors, but if there is a lack of reliability on the Hanna reagents, then I may be switching back.
I have the same issues,been up my doser to fine out the check reagent is off ~1dkh
Going back to salifert
 

Joe Batt

Active Member
View Badges
Joined
Apr 15, 2017
Messages
495
Reaction score
286
Location
Dubai
Rating - 0%
0   0   0
I also notice a difference between batches of reagents. All I do is do the last test with an old reagent followed by the first test with a new reagent. I get an offset that then becomes the baseline for the new bottle. I am more concerned with consistency test to test within a bottle than the difference one to the next.
 

DaveRaz

Active Member
View Badges
Joined
May 14, 2015
Messages
399
Reaction score
280
Location
Memphis TN
Rating - 0%
0   0   0
We went through this a couple years ago and they changed the cap seal from foil to plastic. All was fine until until here recently. I put the checker in the drawer and started using salifert a few weeks ago. The checker is still in the drawer.
 

High pressure shells: Do you look for signs of stress in the invertebrates in your reef tank?

  • I regularly look for signs of invertebrate stress in my reef tank.

    Votes: 15 37.5%
  • I occasionally look for signs of invertebrate stress in my reef tank.

    Votes: 12 30.0%
  • I rarely look for signs of invertebrate stress in my reef tank.

    Votes: 6 15.0%
  • I never look for signs of invertebrate stress in my reef tank.

    Votes: 7 17.5%
  • Other.

    Votes: 0 0.0%