Refractometer Calibration

Aspect

Valuable Member
View Badges
Joined
Apr 8, 2022
Messages
1,562
Reaction score
843
Location
USA
Rating - 100%
2   0   0
When calibrating a refractometer with 35 ppt solution, should the refractometer then ideally be at exactly the 0 line with RODI water or below the 0 line?
 

KrisReef

Last to get paid.
View Badges
Joined
May 15, 2018
Messages
22,081
Reaction score
39,129
Location
Vatican & Las Vegas Penthouse Quarters
Rating - 100%
1   0   0
My understanding is to calibrate the handheld refractometer to 35ppt so that it is reading accurately at the target range. (Use a TDS meter for RO) The refractometer scale is too wide to be accurate across the range from RO to Marine water. If it's set to marine water 35ppt the RO will not be at the zero line, i would expect it to read higher, but whatever direction it will not likely be zero.
 
OP
OP
A

Aspect

Valuable Member
View Badges
Joined
Apr 8, 2022
Messages
1,562
Reaction score
843
Location
USA
Rating - 100%
2   0   0
My understanding is to calibrate the handheld refractometer to 35ppt so that it is reading accurately at the target range. (Use a TDS meter for RO) The refractometer scale is too wide to be accurate across the range from RO to Marine water. If it's set to marine water 35ppt the RO will not be at the zero line, i would expect it to read higher, but whatever direction it will not likely be zero.
It reads below 0 for me
 

penfold2

Active Member
View Badges
Joined
Feb 16, 2013
Messages
239
Reaction score
215
Location
MN
Rating - 0%
0   0   0
I agree with everything the previous poster said. Expect a refractometer to be accurate only at or near the value it's calibrated at. If you want to measure salinity near 0ppt, calibrate with distilled or RODI, measure, then recalibrate with 35ppt sample for future readings near seawater level.
 

SixlineRyan

Active Member
View Badges
Joined
Jun 24, 2019
Messages
107
Reaction score
86
Location
Cincinnati
Rating - 0%
0   0   0
What ever you do, it's worth it to make extra sure you've got it right. I borked mine and ended up with salinity that was waaay to high.
 

KrisReef

Last to get paid.
View Badges
Joined
May 15, 2018
Messages
22,081
Reaction score
39,129
Location
Vatican & Las Vegas Penthouse Quarters
Rating - 100%
1   0   0
It reads below 0 for me
Yeah, light refraction measurements use physics principals and light bending between two mediums to calculate the results. I am too lazy to look up the physics and the results, I just remember that calibration for hobby grade equipment is best done at the target level.

Side bar: I just purchased a precision glass floating hydrometer that is miles away from being accurate. I have a plastic swingarm that appears to be more accurate. Getting a good tool that repeats measurement at the target level is what we need for the hobby. I bought the expensive floating hydrometer to see how accurate my hydrometer is and was shocked how poorly the hydrometer is.

I would recommend that you rinse with RO and dry a refractometer after you use it and put it away to reduce rusting of the metal parts.
 
OP
OP
A

Aspect

Valuable Member
View Badges
Joined
Apr 8, 2022
Messages
1,562
Reaction score
843
Location
USA
Rating - 100%
2   0   0
Yeah, light refraction measurements use physics principals and light bending between two mediums to calculate the results. I am too lazy to look up the physics and the results, I just remember that calibration for hobby grade equipment is best done at the target level.

Side bar: I just purchased a precision glass floating hydrometer that is miles away from being accurate. I have a plastic swingarm that appears to be more accurate. Getting a good tool that repeats measurement at the target level is what we need for the hobby. I bought the expensive floating hydrometer to see how accurate my hydrometer is and was shocked how poorly the hydrometer is.

I would recommend that you rinse with RO and dry a refractometer after you use it and put it away to reduce rusting of the metal parts.
Yeah I am never really sure what my actual salinity is, I have 3 pre calibrated solutions that all give me different readings. I am most confident about the brightwell solution which I just try to keep my salinity always stable referring to the calibration with that solution. I just wanted to see if it would measure accurately at rodi levels which looks like it doesn't. I'm waiting back on an icp to see my true salinity level.
 

KrisReef

Last to get paid.
View Badges
Joined
May 15, 2018
Messages
22,081
Reaction score
39,129
Location
Vatican & Las Vegas Penthouse Quarters
Rating - 100%
1   0   0
In case you have not seen this?

Randy provides instruction on how to use table salt to make calibration solutions for different refractometer, hydrometer, electronic meters, in this article. Very helpful to know that table salt mixtures of different concentrations are used to calibrate different meters. Turns out "calibration solutions" are not universally useful for all instruments that need to be calibrated. HTH.
 

NanoSteam

ANTI 100% Blues Only Club
View Badges
Joined
Dec 23, 2019
Messages
2,887
Reaction score
13,341
Location
SoCal
Rating - 0%
0   0   0
Yeah, light refraction measurements use physics principals and light bending between two mediums to calculate the results. I am too lazy to look up the physics and the results, I just remember that calibration for hobby grade equipment is best done at the target level.

Side bar: I just purchased a precision glass floating hydrometer that is miles away from being accurate. I have a plastic swingarm that appears to be more accurate. Getting a good tool that repeats measurement at the target level is what we need for the hobby. I bought the expensive floating hydrometer to see how accurate my hydrometer is and was shocked how poorly the hydrometer is.

I would recommend that you rinse with RO and dry a refractometer after you use it and put it away to reduce rusting of the metal parts.

Curious which glass Hydrometer did you pick up? The TM one is supposed to be the gold standard. I also have a more fancy refractometer designed for seawater (Vee-Gee STX-3), those are supposed to be calibrated with RODI according to the manufacture.
 

GSPClown94

Well-Known Member
View Badges
Joined
Jan 17, 2018
Messages
658
Reaction score
661
Location
VA
Rating - 0%
0   0   0
In case you have not seen this?

Randy provides instruction on how to use table salt to make calibration solutions for different refractometer, hydrometer, electronic meters, in this article. Very helpful to know that table salt mixtures of different concentrations are used to calibrate different meters. Turns out "calibration solutions" are not universally useful for all instruments that need to be calibrated. HTH.
This! I only trust Randy's recipe to make a calibration solution. Too many times i have found those commercially available solutions to be way off.
 

mcarroll

10K Club member
View Badges
Joined
Jan 8, 2012
Messages
15,213
Reaction score
8,968
Location
Virginia
Rating - 0%
0   0   0
When calibrating a refractometer with 35 ppt solution, should the refractometer then ideally be at exactly the 0 line with RODI water or below the 0 line?
It's a minor difference either way if you're following the directions and calibrating properly.

The only situation I know where ultimate precision is required is if you're trying to do hyposalinity for new fish. In that case you're trying to make the water as low-salinity as the fish will tolerate before it causes them damage. So, being accurate is crucial.

But for general tank care, mixing saltwater, etc. I'm not sure it matters at all.

From the RHF article linked in GSPClown's post:
It is widely believed that only pure water is required to calibrate refractometers. That fact is true of some refractometers [....] and it may be acceptable for routine calibration[...]

Edited just for brevity. Emphasis mine. Please read the article if you haven't already.

Again: Pure water (even tap water, in my experience) is totally acceptable for routine calibration.
 

dedragon

5000 Club Member
View Badges
Joined
Feb 23, 2019
Messages
6,901
Reaction score
4,939
Rating - 0%
0   0   0
i use the tm hydrometer to calibrate if i need to test multiple samples. if it only for one or two readings, i just use the tm and no refractometer
 

dedragon

5000 Club Member
View Badges
Joined
Feb 23, 2019
Messages
6,901
Reaction score
4,939
Rating - 0%
0   0   0
when calibrating with 35ppt, i wouldnt worry about what rodi was reading at all
 

KrisReef

Last to get paid.
View Badges
Joined
May 15, 2018
Messages
22,081
Reaction score
39,129
Location
Vatican & Las Vegas Penthouse Quarters
Rating - 100%
1   0   0
Curious which glass Hydrometer did you pick up? The TM one is supposed to be the gold standard. I also have a more fancy refractometer designed for seawater (Vee-Gee STX-3), those are supposed to be calibrated with RODI according to the manufacture.
That’s why I bought the TM, but I might have gotten a dud?
IMG_1810.jpeg

It’s too buoyant- seen here floating in Randy’s Morton’s Iodized salt calibration solution. I raised the temperature to 30C and it was still floating too high. 1.026 +
 

NanoSteam

ANTI 100% Blues Only Club
View Badges
Joined
Dec 23, 2019
Messages
2,887
Reaction score
13,341
Location
SoCal
Rating - 0%
0   0   0
Curious which glass Hydrometer did you pick up? The TM one is supposed to be the gold standard. I also have a more fancy refractometer designed for seawater (Vee-Gee STX-3), those are supposed to be calibrated with RODI according to the manufacture.
That’s why I bought the TM, but I might have gotten a dud?
IMG_1810.jpeg

It’s too buoyant- seen here floating in Randy’s Morton’s Iodized salt calibration solution. I raised the temperature to 30C and it was still floating too high. 1.026 +
Interesting this is the first complaint Ive heard of it. Mine matches with my refractometer so I havent questioned it but I've been itching to get a swing arm for old times sake and simplicity sake. Which one are you using at the moment?
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,232
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
When calibrating a refractometer with 35 ppt solution, should the refractometer then ideally be at exactly the 0 line with RODI water or below the 0 line?

Many refractometers used in the hobby are brine refractometers repurposed from other industries to sell to reef hobbyists. I suspect many resellers do not even understand what they are selling and why it is not ideal.

A brine refractometer is made to measure salinity of sodium chloride, which is a lot, but not all of what seawater is.

A true seawater refractometer is designed to measure the salinity of seawater.

A perfectly made and perfectly calibrated brine refractometer cannot read both zero ppt and 35 ppt correctly. One must be off, as you have found. If you calibrated with an accurate standard at 35 ppt, then it will read 35 ppt correctly, and will even do so if less than perfectly made.

A perfectly made and perfectly calibrated true seawater refractometer can read both zero ppt and 35 ppt correctly. If you calibrated with an accurate standard at 35 ppt, then it will read 35 ppt correctly, and will even do so if less than perfectly made.

Thus, there are several scenarios where calibration with 0 ppt solutions can lead to errors, and no scenarios where calibration with an accurate 35 ppt (or similar salinity) standard will lead to errors.

That said, commercial salinity standards seem to have a poor track record of accuracy for reasons that I cannot understand. They are cheap and easy to make. Why do they go wrong?

For this reason, I use and recommend a diy using a good scale for both the sodium chloride and the water. Using a salt mix to make a standard is also not a good plan since it can have variable amounts of water in the solids.
 

TOP 10 Trending Threads

WHAT AMOUNT OF LIVE ROCK AND SAND SHOULD BE PRIORITIZED FOR OPTIMAL BIODIVERSITY/FILTRATION?

  • 100% live rock + bagged sand

    Votes: 34 27.6%
  • 100% dry rock + 100% live sand

    Votes: 43 35.0%
  • 50/50 live/dry rock, 50/50 live/bagged sand

    Votes: 27 22.0%
  • 75% live rock, 25% live sand

    Votes: 11 8.9%
  • 25% live rock, 75% live sand

    Votes: 8 6.5%
Back
Top