Hey there,

I'm using one of those basic refractometers that claim to adjust their scale based on ambient/sample temperatures (ATC). I've found that it holds calibration extremely well, but I still check it every other test.

One thing I haven't seen being discussed is my particular method of calibrating. I set my tank at 77F (checked with multiple thermometers) and float the calibration solution AND the refractometer prism (where the sample is placed) to bring them both to the same temp.

Most people say to just float the solution, put it on the prism, and wait 30 seconds for the temp to 'stabilize' before reading. However, this doesn't make much sense to me, as if ambient temperatures are several degrees cooler than the sample water (~68F in my case) the water will cool below the target 77F. A couple drops of water won't bring the prism and surrounding aluminum to 77 degrees.

In my experience, my method and the "wait 30 seconds" method provides different results - floating the prism to temp makes the reading about 1 PPT lower as I expect. Does this mean the ATC actually isn't that accurate?

So, which method would be best? I don't have a true glass hydrometer to test this with. I think by ensuring both the refractometer and the solution are both at 77F, it would be closer to a TRUE 77 degrees rather than somewhere between room temp and 77.
 
Last edited:

exnisstech

Grumpy old man
View Badges
Joined
Feb 11, 2019
Messages
18,920
Reaction score
30,366
Location
Ashland Ohio
Rating - 0%
0   0   0
I use a TM hydrometer to test fresh mixed water at 77 degrees adjusted to 1.025 salinty then use that same water to calibrate my refractometers.
I've tested salt water at around 60 degrees using a ATC refractometer which is what my basement temp is and again when the water is warmed up and I get the same reading at both temps so I'm not sure warming everything up is necessary but I suppose it can't hurt.
 
OP
OP
SeansNanoReef

SeansNanoReef

Active Member
View Badges
Joined
Jul 4, 2019
Messages
117
Reaction score
93
Rating - 0%
0   0   0
Interesting, sometimes I get up to a 2 ppt difference with mine. Maybe my ATC function is defective or something.
 

fish_collector

Valuable Member
View Badges
Joined
Jul 24, 2021
Messages
1,292
Reaction score
2,671
Location
USA
Rating - 0%
0   0   0
Your salinity isn’t critical enough to have to go through all of that. Somewhere near 35 ppt as measured with your refractometer is good.

For 20 years I used an instant ocean swing arm to measure salinity, it worked. Nowadays everyone fusses over how to really measure it, but in the end it’s not that critical.

If you are in need of a true hyposalinity treatment where the salinity is just barely enough to not kill your fish but eradicate a pathogen, then yes it may be important to be able to measure it accurately, but otherwise just roll with what your refractometer says on the fly and don’t sweat it, it’s good enough.

FWIW, I’ve compared my cheap Amazon refractometer with my buddies $350 Milwaukee unit and they read exactly the same. I’ve never messed with the adjusting screw.
 
OP
OP
SeansNanoReef

SeansNanoReef

Active Member
View Badges
Joined
Jul 4, 2019
Messages
117
Reaction score
93
Rating - 0%
0   0   0
I totally agree, one doesn't really need to know exactly what their salinity is. I'm being a little critical of it now because I was using one of those conductivity style pens that turned out to be completely inaccurate even when calibrated, and I was dealing with mysterious invert deaths for a little while. Turns out my salinity was actually 40ppt. That left a bad taste so I guess I'm going a bit overboard with my attention to salinity, specifically accuracy.
 

fish_collector

Valuable Member
View Badges
Joined
Jul 24, 2021
Messages
1,292
Reaction score
2,671
Location
USA
Rating - 0%
0   0   0
Ok fair enough. The Topic Marin hydrometer that exnisstech referred to is the gold standard for something to compare to. I understand your position with it all, 40 probably wasn’t high enough to kill things though. I’ve let my ATO go dry and end up with 40 in my SPS tank, and the only reason I checked the salinity was to find out how much fresh water I had to add 🤪
 
OP
OP
SeansNanoReef

SeansNanoReef

Active Member
View Badges
Joined
Jul 4, 2019
Messages
117
Reaction score
93
Rating - 0%
0   0   0
Ok fair enough. The Topic Marin hydrometer that exnisstech referred to is the gold standard for something to compare to. I understand your position with it all, 40 probably wasn’t high enough to kill things though. I’ve let my ATO go dry and end up with 40 in my SPS tank, and the only reason I checked the salinity was to find out how much fresh water I had to add 🤪
Yeah, I didn't have any issues with my fish or corals, just snails...I think it was because every other element was also through the roof due to my particular salt, and I didn't "acclimate." As soon as I lowered my salinity they have been doing great.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Hey there,

I'm using one of those basic refractometers that claim to adjust their scale based on ambient/sample temperatures (ATC). I've found that it holds calibration extremely well, but I still check it every other test.

One thing I haven't seen being discussed is my particular method of calibrating. I set my tank at 77F (checked with multiple thermometers) and float the calibration solution AND the refractometer prism (where the sample is placed) to bring them both to the same temp.

Most people say to just float the solution, put it on the prism, and wait 30 seconds for the temp to 'stabilize' before reading. However, this doesn't make much sense to me, as if ambient temperatures are several degrees cooler than the sample water (~68F in my case) the water will cool below the target 77F. A couple drops of water won't bring the prism and surrounding aluminum to 77 degrees.

In my experience, my method and the "wait 30 seconds" method provides different results - floating the prism to temp makes the reading about 1 PPT lower as I expect. Does this mean the ATC actually isn't that accurate?

So, which method would be best? I don't have a true glass hydrometer to test this with. I think by ensuring both the refractometer and the solution are both at 77F, it would be closer to a TRUE 77 degrees rather than somewhere between room temp and 77.

With an ATC refractometer, the temp of a standard and the temp
Of the sample should not matter, at least within its range of ATC effectiveness. If you do see temp differences, then yes, it means the correction is not perfect and having tank and 35 ppt standard at close to the same temp is best.

Realize that drops on the refractometer fairly quickly come to the temp of the refractometer, not the starting water temp.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Ok fair enough. The Topic Marin hydrometer that exnisstech referred to is the gold standard for something to compare to. I understand your position with it all, 40 probably wasn’t high enough to kill things though. I’ve let my ATO go dry and end up with 40 in my SPS tank, and the only reason I checked the salinity was to find out how much fresh water I had to add 🤪

Just to clarify, it may be very convenient and perfectly adequate for reefers to do that, but it is not meeting the definition of a gold standard. Conductivity is truly the gold standard,,and is how salinity is actually defined these days.
 

rtparty

7500 Club Member
View Badges
Joined
Jan 19, 2010
Messages
8,666
Reaction score
14,872
Location
Utah
Rating - 100%
4   0   0
Salinity is absolutely essential and the most important parameter to get correct. All elements are predicated upon a correct salinity to begin with.

We see it almost daily. Someone will come on claiming a bad batch of salt because the parameters are super low or super high but the reality is their salinity is wrong.

They will claim their “tank has always run at low levels” but in reality their salinity was incorrect.

Anyone saying otherwise is ignoring basic science
 

Waters

"...in perfect isolation, here behind my wall."
View Badges
Joined
Nov 5, 2013
Messages
9,346
Reaction score
20,852
Location
Mentor, OH
Rating - 0%
0   0   0
I haven't calibrated my refractometer since the day I got it, 4 years ago lol. It still reads identical to my Milwaukee. If either shows something way off, then I will dig a bit further.
 

BryanM

7500 Club Member
View Badges
Joined
Jun 27, 2024
Messages
7,651
Reaction score
9,361
Location
Morgan Hill
Rating - 100%
1   0   0
Hey there,

I'm using one of those basic refractometers that claim to adjust their scale based on ambient/sample temperatures (ATC). I've found that it holds calibration extremely well, but I still check it every other test.

One thing I haven't seen being discussed is my particular method of calibrating. I set my tank at 77F (checked with multiple thermometers) and float the calibration solution AND the refractometer prism (where the sample is placed) to bring them both to the same temp.

Most people say to just float the solution, put it on the prism, and wait 30 seconds for the temp to 'stabilize' before reading. However, this doesn't make much sense to me, as if ambient temperatures are several degrees cooler than the sample water (~68F in my case) the water will cool below the target 77F. A couple drops of water won't bring the prism and surrounding aluminum to 77 degrees.

In my experience, my method and the "wait 30 seconds" method provides different results - floating the prism to temp makes the reading about 1 PPT lower as I expect. Does this mean the ATC actually isn't that accurate?

So, which method would be best? I don't have a true glass hydrometer to test this with. I think by ensuring both the refractometer and the solution are both at 77F, it would be closer to a TRUE 77 degrees rather than somewhere between room temp and 77.
This is just another reason I switched to the TM Hydrometer. No fussing around.
 

GSPClown94

Well-Known Member
View Badges
Joined
Jan 17, 2018
Messages
658
Reaction score
661
Location
VA
Rating - 0%
0   0   0
I also have experienced this where the temperature of the actual refractometer changes the reading, I believe this is from something broken within the refractometer related to the automatic temperature control. I would look into getting a new device to measure salinity.
 

fish_collector

Valuable Member
View Badges
Joined
Jul 24, 2021
Messages
1,292
Reaction score
2,671
Location
USA
Rating - 0%
0   0   0
Just to clarify, it may be very convenient and perfectly adequate for reefers to do that, but it is not meeting the definition of a gold standard. Conductivity is truly the gold standard,,and is how salinity is actually defined these days.
Yes I mean from a hobbyists perspective.
 
OP
OP
SeansNanoReef

SeansNanoReef

Active Member
View Badges
Joined
Jul 4, 2019
Messages
117
Reaction score
93
Rating - 0%
0   0   0
I'm leaning towards a true seawater refractometer like the VeeGee. I have a small tank so filling a 1000 ml graduated cylinder for a hydrometer every time would be just as big or more of a hassle for my situation.

For now my temperature equalization method seems to be effective, the cheap refractometer hasn't needed any adjustments since purchase.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Yes I mean from a hobbyists perspective.
It's not worth a debate, and the TM hydrometer is generally a fine instrument that I'm not going to criticize if one validates its accuracy with a standard, but many hobbyists and Reef2Reef members do use high quality instruments to measure conductivity. :)
 

daikaijureefer

Active Member
View Badges
Joined
Oct 4, 2025
Messages
144
Reaction score
114
Location
Atlanta, GA
Rating - 0%
0   0   0
Just to clarify, it may be very convenient and perfectly adequate for reefers to do that, but it is not meeting the definition of a gold standard. Conductivity is truly the gold standard,,and is how salinity is actually defined these days.
Randy, can you recommend a conductivity probe/ meter that meets these qualifications?
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,230
Reaction score
92,241
Location
Massachusetts, United States
Rating - 0%
0   0   0
Randy, can you recommend a conductivity probe/ meter that meets these qualifications?


Since they are quite expensive, many hobbyists get them on ebay.

For decades, I used an Orion Model 128 with a four electrode probe. Many old Orions are good. There are different probe models, and some work in seawater, while others are optimized for lower conductivity. It finally broke and I bought this very old one from VWR on ebay:


Rishma bought this meter:

 

fish_collector

Valuable Member
View Badges
Joined
Jul 24, 2021
Messages
1,292
Reaction score
2,671
Location
USA
Rating - 0%
0   0   0
It's not worth a debate, and the TM hydrometer is generally a fine instrument that I'm not going to criticize if one validates its accuracy with a standard, but many hobbyists and Reef2Reef members do use high quality instruments to measure conductivity. :)
No debate at all Randy. Thank you for mentioning there was something better.
 

TOP 10 Trending Threads

WHAT AMOUNT OF LIVE ROCK AND SAND SHOULD BE PRIORITIZED FOR OPTIMAL BIODIVERSITY/FILTRATION?

  • 100% live rock + bagged sand

    Votes: 34 27.9%
  • 100% dry rock + 100% live sand

    Votes: 42 34.4%
  • 50/50 live/dry rock, 50/50 live/bagged sand

    Votes: 27 22.1%
  • 75% live rock, 25% live sand

    Votes: 11 9.0%
  • 25% live rock, 75% live sand

    Votes: 8 6.6%
Back
Top