Can’t seem to get my salinity probe calibrated properly. Left it out to dry all night. Can’t get past first part of auto cal. Did manual cal. And it’s closer but still reading 42 when I know it’s 35. I have tc factor set at 2.2. Getting frustrated. It’s new btw.
 

mckinney0171

Well-Known Member
View Badges
Joined
Jul 10, 2017
Messages
554
Reaction score
627
Location
Richmond Hill, GA
Rating - 0%
0   0   0
make sure you are shaking it real good inside the calibration solution bag to get any bubbles out. that was the issue that i had. once you get these things calibrated they are pretty bulletproof. i honestly have never had to re-calibrate mine even though i probably should. i also keep mine in the water sideways and away from the flow of current in my sump to keep bubbles from getting trapped underneath. not sure if that helps or not but that's what i do.
 

SPR1968

No, it wasn’t expensive dear....
View Badges
Joined
Feb 21, 2017
Messages
20,109
Reaction score
119,310
Location
Nottinghamshire England
Rating - 0%
0   0   0
I had the same problem especially when using the solution.

So when I calibrate I do it when my Refractometer (which I trust because everything is alive!)reads 35 and then when the calibration calls for the probe to go in the solution, it goes back into the tank water instead. And it’s generally been fine doing this.
 
Last edited:
OP
OP
Kzguns

Kzguns

Active Member
View Badges
Joined
Jan 9, 2018
Messages
191
Reaction score
134
Rating - 0%
0   0   0
make sure you are shaking it real good inside the calibration solution bag to get any bubbles out. that was the issue that i had. once you get these things calibrated they are pretty bulletproof. i honestly have never had to re-calibrate mine even though i probably should. i also keep mine in the water sideways and away from the flow of current in my sump to keep bubbles from getting trapped underneath. not sure if that helps or not but that's what i do.

Ya I shook it and brought temp up etc. I’ve tried several times. Probably have to get in touch with Neptune on Monday if nothing changes
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0
Can’t seem to get my salinity probe calibrated properly. Left it out to dry all night. Can’t get past first part of auto cal. Did manual cal. And it’s closer but still reading 42 when I know it’s 35. I have tc factor set at 2.2. Getting frustrated. It’s new btw.

That means the calibration solution is reading 42 ppt when it is a 35 ppt calibration solution, or are you talking about tank water reading 42 ppt after calibration?
 
OP
OP
Kzguns

Kzguns

Active Member
View Badges
Joined
Jan 9, 2018
Messages
191
Reaction score
134
Rating - 0%
0   0   0
That means the calibration solution is reading 42 ppt when it is a 35 ppt calibration solution, or are you talking about tank water reading 42 ppt after calibration?

Actually I haven’t been able to get it to read correctly in the solution or my tank. I ordered more solution so hopefully I can get it sorted.
 
OP
OP
Kzguns

Kzguns

Active Member
View Badges
Joined
Jan 9, 2018
Messages
191
Reaction score
134
Rating - 0%
0   0   0
I had the same problem especially when using the solution.

So when I calibrate I do it when my Refractometer (which I trust because everything is alive!)reads 35 and then when the calibration calls for the probe to go in the solution, it goes back into the tank water instead. And it’s generally been fine doing this.

I’ll give that a shot.
 
U

User1

Guest
View Badges
Similar issue here. Upon first calibration I used the shipped solution and it was off. Read a bit on Neptunes forums and found a thread about how to properly or at least consistently, calibrate the probe. One piece of information that I missed was to put the calibration packet / fluid into the tank or sump to bring it up to the same temperature. Then like others mentioned move it around inside the solution to remove any bubbles. Once I did this it was reading more or less the same has my digital tester.

So for me the main key was to bring it to the same temperature has the tank then calibrate.
 

Artillerydrill

Active Member
View Badges
Joined
May 25, 2017
Messages
251
Reaction score
176
Location
EL PASO
Rating - 0%
0   0   0
Get some canned air to dry it out that should get you to zero at least. Then throw it in the tank for your 35 reading. As long as you know what the numbers are you will know if you are getting swings and the ato fails etc.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0
Get some canned air to dry it out that should get you to zero at least. Then throw it in the tank for your 35 reading. As long as you know what the numbers are you will know if you are getting swings and the ato fails etc.

Why do you need to dry it out? That's not a necessary part of calibrating a normal conductivity meter. :)
 

JVU

Active Member
View Badges
Joined
Mar 14, 2017
Messages
173
Reaction score
132
Rating - 0%
0   0   0
It’s part of the Apex calibration routine, presumably to set the zero point.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0
Doesn't it read zero before being "dried"?

Maybe it's some unusual design that doesn't drain itself well, but all the conductivity meters I've ever used (many over the years, hobby to high end) drop immediately to zero conductivity immediately on being withdrawn from the solution.

Presumably they are worried about conductivity through the residual water stuck to the sides of the device.

In any case, there is typically no "setting" of the zero point. Zero is zero.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0
FWIW, it is really shocking to me how many people have issues with controllers and conductivity. I have to believe they just have a problematic implementation somehow, since such issues do not seem to arise nearly as often in using stand alone conductivity meters.

I even just watched a long video by them discussing the "issues". Why are there so many issues? lol
 

JVU

Active Member
View Badges
Joined
Mar 14, 2017
Messages
173
Reaction score
132
Rating - 0%
0   0   0
As I understand it is the residual water issue allowing some conductivity. People have reported that it won’t calibrate unless actively dried during the step in calibration that calls for taking it out of solution. There are more complaints and frustration with the Apex conductivity probe than all the others combined as far as I’ve seen.
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0

Artillerydrill

Active Member
View Badges
Joined
May 25, 2017
Messages
251
Reaction score
176
Location
EL PASO
Rating - 0%
0   0   0
Mine doesn’t drop to zero and when it can’t hit zero it can’t be properly calibrated. Of course you can use simple math to figure the deviation in the numbers but what’s the point if you’ve spent all the money
 

Randy Holmes-Farley

Reef Chemist
View Badges
Joined
Sep 5, 2014
Messages
89,791
Reaction score
93,380
Location
Massachusetts, United States
Rating - 0%
0   0   0
Mine doesn’t drop to zero and when it can’t hit zero it can’t be properly calibrated. Of course you can use simple math to figure the deviation in the numbers but what’s the point if you’ve spent all the money

Then it must either be a bad probe design, or bad software. :(

I just tried my Orion model 128 conductivity meter with a 4 electrode probe. I dropped it in a salt solution that was about 25 mS/cm. It gives a new reading about once a second, and with one update after being pulled from the solution, it read 0.0 mS/cm. I switched it to a lower conductivity scale, and it read about 50 uS/cm. So the drop within a few seconds is a factor of 500 in terms of conductivity (and that's a salty water solution on it, not residual RO/DI).

You'd think if they used a good probe and good software, you would not need to dry it.
 
OP
OP
Kzguns

Kzguns

Active Member
View Badges
Joined
Jan 9, 2018
Messages
191
Reaction score
134
Rating - 0%
0   0   0
Seems to be stabilizing now. Hmmmm
cfdc5c5a67710c5157fa9e87aabe4897.jpg
 

Vincent100

Valuable Member
View Badges
Joined
Jul 5, 2016
Messages
1,423
Reaction score
1,820
Location
Kent
Rating - 0%
0   0   0
I've never been able to get mine working right ......don't even take any notice of it anymore , waste of space lol

Screenshot_20180402-220007.png


Screenshot_20180202-231226.png


Screenshot_20180116-162435.png
 

TOP 10 Trending Threads

DO YOU THINK TECHNOLOGICAL ADVANCEMENTS ARE MORE HELPFUL OR HURTFUL TO REEFING?

  • More helpful.

    Votes: 11 39.3%
  • More hurtful.

    Votes: 2 7.1%
  • I think it depends mostly on the technology.

    Votes: 11 39.3%
  • I think it dependsmostly on the reefer behind the technology.

    Votes: 9 32.1%
Back
Top
Home
Post thread…
Market
What's new