New pH electrode reading and calibration issue

Users Who Are Viewing This Thread (Total: 1, Members: 0, Guests: 1)

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
I just received my Profilux 4e starter set with temp and pH probe.
When I calibrate the pH electrode, the ADC reading is always between 475-477, regardless whether the probe is in pH4.0, 7.0, or 8.2 fluid.
I believe that is not correct, and is causing the pH reading after the calibration to be nonsense (10.50 or 3.00).
When I change the measuring data in GHL controller for this sensor from pH to ORB/Redux and back to pH, the pH reading is around 6.78, whatever liquid the probe is put into.
Is this a faulty electrode? How could one check that the tip is ok?
 

Vinny@GHLUSA

Valuable Member
View Badges
Joined
Jul 27, 2016
Messages
1,148
Reaction score
1,507
Rating - 0%
0   0   0
Exactly which port is this probe connected to?
What is this probe port currently set to? pH or redox mode?
 
OP
OP
Kobooi

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
Exactly which port is this probe connected to?
What is this probe port currently set to? pH or redox mode?
It is connected to the pH/Redox port directly.
The port is set to pH. After calibration it shows pH 3.0 or pH 10.5. When I change the port setting to redox, and then back to pH, it looks like the calibration data is deleted because then it shows around 6.77 pH, which is what it showed before the first calibration.
 

Vinny@GHLUSA

Valuable Member
View Badges
Joined
Jul 27, 2016
Messages
1,148
Reaction score
1,507
Rating - 0%
0   0   0
@Kobooi
If this is still an issue please open a support ticket for direct assistance.
Provide a copy of your sales receipt and serial number of your device.
 
OP
OP
Kobooi

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
I did open a ticket with GHL last Saturday.
With no response on the ticket by Wednesday evening, I opened another ticket.
No response yet.
Based on the feedback and reviews of GHL support, I had higher expectations...
 

Vinny@GHLUSA

Valuable Member
View Badges
Joined
Jul 27, 2016
Messages
1,148
Reaction score
1,507
Rating - 0%
0   0   0
Ok good.
At times, it may take up to 2 working days to receive a reply.
 
OP
OP
Kobooi

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
Understood, but it's end of business day 5 now, with no response.
Ticket number is #2912
 
OP
OP
Kobooi

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
Business day 6 of my ticket, with still no response from GHL.
The auto-notification email from GHL says the aim is to respond in 1 business day, but to allow up to 3 business days if very busy.
So far I'm not impressed with GHL. A dead-on-arrival and no response for 6 business days already.
 

ajhudson15

Valuable Member
View Badges
Joined
Mar 26, 2017
Messages
1,307
Reaction score
980
Location
Bloomington, IN
Rating - 0%
0   0   0
GHL has picked up the ticket and are making good. Phew!
may I ask what you found out. I had a ph probe be fine and then one day when I unplugged everything to move it it now reads 10.50 all the time and wont change which I know is wrong because its in my cal reactor
 
OP
OP
Kobooi

Kobooi

Community Member
View Badges
Joined
Aug 5, 2020
Messages
44
Reaction score
65
Location
Belgium
Rating - 0%
0   0   0
GHL sent me a new probe, which worked. I didn't see anything different on the old one compared to the new one. Something must just have been faulty on the old one. I threw it away immediately after I received a working new one.
 

ajhudson15

Valuable Member
View Badges
Joined
Mar 26, 2017
Messages
1,307
Reaction score
980
Location
Bloomington, IN
Rating - 0%
0   0   0
GHL sent me a new probe, which worked. I didn't see anything different on the old one compared to the new one. Something must just have been faulty on the old one. I threw it away immediately after I received a working new one.
that was my concern. I bought a brand new brs ph probe because I read that they work to and it did for about a year and all of a sudden when I disconnected it and reconnected it it stopped working.
 

Lasse

10K Club member
View Badges
Joined
Mar 20, 2016
Messages
10,861
Reaction score
29,839
Location
Källarliden 14 D Bohus, Sweden
Rating - 0%
0   0   0
Normally you can see that a probe is going to break down when you calibrate. When the difference between ADC1 and ADC 2 differ to much or to little and ADC2 (in this case) going near 0 - its time to replace the electrode.



1611773531463.png


Sincerely Lasse
 